> ultrasound hi vision 6500 problem

ultrasound hi vision 6500 problem

(26)
(2)

M.O.A.

ultrasound hi vision 6500 problem

July 14, 2017 01:30

Dear s
I have Hitachi HI-Vision 6500, the following error massage appears
(Getting System ID has been failed)
And the system can’t see any probe
Anybody can help
Regards

Log inor Register

to rate and post a comment

(2)
(1)

Sara Khan

re: ultrasound hi vision 6500 problem

July 16, 2017 03:41

Have the same problem with Hitachi eub 6500

Log inor Register

to rate and post a comment

(61)
(8)

Uldis Verdins

re: ultrasound hi vision 6500 problem

July 17, 2017 03:27

Hi!
Start repair process with checking system time.
If BIOS battery is gone, system time is bad and system ID isn`t correct.
Good lack,
Uldis

Log inor Register

to rate and post a comment

(26)
(2)

M.O.A.

re: re: ultrasound hi vision 6500 problem

July 18, 2017 08:15

many thanks for you dear
yes the bios battery was bad
I replace it
and adjust the time
what your suggestion for next step
regards

Log inor Register

to rate and post a comment

(61)
(8)

Uldis Verdins

re: re: re: ultrasound hi vision 6500 problem

July 18, 2017 10:27

Next. You must check system HDD for surface error.
If You`l find sam BAD sector, this is problem root.
For checking HDD You must take out it and connect to Windows XP or Linux PC and check it with Cristal HDD, Hddscan programm or similar. Please do not connect this HDD to Windovs Vista, 7 or 10, becose this system will bad situation increase bigger (in case if HDD has Bad Block).
God idea is booting (regular PC or this Ultrasound) from live CD with similar checking tools.
Good lack,
Uldis

Log inor Register

to rate and post a comment