Siremobil 2000

September 09, 2017 11:16

Hi
For C-Arm Siemens Siremobil 2000, what is
err 0021 and how to fix
Also I need the Service manual and trouble shooting.
Thanks in advance.

Log inor Register

to rate and post a comment

(208)
(27)
(27)

Wayne Horsman

re: Siremobil 2000

September 11, 2017 02:58

0021 I.I. power supply error
Cause : Message from I.I. power supply. The power supply for the image intensifier (high voltage) is out-of-tolerance.
Active: Constantly
Possible trouble / help : I.I. power supply
. Interruption in indication lead from I.I. to board D6; operating voltage missing (also triggers error 0020).

email me if you need a replacement poor supply.
wayne@columbiaimaging.com

Log inor Register

to rate and post a comment

(3)

Alaa Abd elLateef

re: re: Siremobil 2000

September 11, 2017 04:51

Thanks alot for your kind reply...
I already fixed the problem and now every thing is OK..
Many thanks again
Alaa Mohammed

Log inor Register

to rate and post a comment

(208)
(27)
(27)

Wayne Horsman

re: re: re: Siremobil 2000

September 12, 2017 09:20

for my benefit next time this comes up, what did you find?

Log inor Register

to rate and post a comment

(3)

Alaa Abd elLateef

re: re: re: re: Siremobil 2000

September 13, 2017 09:06

I got the Service manual few hours before going for repaire;so I followed the error led indicators in D11;it was OK.. the green led.
So I checked the led in D120; the green led wasn't light after the self test...I checked all connections...the socket which feeds the board was loosen..so I fixed it and every thing became OK
Thanks