Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 14, 2019 04:37

Hi, dotmed stuff!
The device starts - fine. But when the X-ray process begins, the following error appears on the display:
"Fault
Measured kV UNDER"
Photo: https://ibb.co/6Pn1GQG
Nowhere about this error is not written. Board checking everything is fine.
Tell me please exactly what this error is connected with?

Log inor Register

to rate and post a comment

(48)
(4)

Emrah Kambak

re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:06

Hi. My name is emrah. Please check inverter fuse. 100A. If blown check igbt.

Log inor Register

to rate and post a comment

(2)

Insendio fx

re: re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 10:59

Hi, Emrah. We have already checked the inverter fuse - no defective inverters were found. Tomorrow check igbt. Thanks for your reply.

Log inor Register

to rate and post a comment

(39)
(3)
(2)

Alessandro Lazzari

re: Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:06

This error means that kv feedback is

lower then expected. Measure by
oscilloscope kvfb+ and - and compare
with the set output valute to evacuate where is the fault

Log inor Register

to rate and post a comment

(48)
(4)

Emrah Kambak

re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:07

Please contact me ekambak@gmail.com

Log inor Register

to rate and post a comment

(39)
(3)
(2)

Alessandro Lazzari

re: Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:08

Hi ! I see this error before. You have to check kv+/- feedback that comes from HV tank and compare to real Kv setting measured then you can judge and try to adjust the lower value or continue troubleshooting on Hv control-Hv-Tube

Log inor Register

to rate and post a comment

(519)
(45)

DMEDS

re: Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 22, 2019 10:44

i have all manuals and service software,
if you need any contact me,
magmesme@gmail.com
thanks
regards

Log inor Register

to rate and post a comment

Laser-Tech.com: Used Lasers & IPL - Buy, Sales, Repair, Parts 855-LASER-00

Are you looking to Buy or Sell a Laser? Do you need Service, Repair or Parts for you Laser or IPL system? Call LASER-TECH today at 855-527-3700 We Sell and Service all major brands and manufactures!

 

advertisement

 

advertisement

 

advertisement

 

advertisement
对这个站点的通入和用途是受期限和条件我们支配 法律公告 & 保密性通知
物产和业主对 DOTmed.com,公司 Copyright ©2001-2019 DOTmed.com, Inc.
版权所有