GE DMR Plus Error E04 and E71 Assistance

Tags: Mammo, GE

(1)
(1)

David Burberry

GE DMR Plus Error E04 and E71 Assistance

April 15, 2019 10:31

Good day,

We have a GE Seno DMR+ mammography unit giving the following errors:

E04 - Arm Fatal Failure - 027/002

E71 - Bucky Failure - 118/012

Any advise in resolving this would be greatly appreciated.

Thanks

1 person liked this
Log inor Register

to rate and post a comment

(43)
(6)

Francesco Didonato

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:05

Hello gli have problem on Grid.
Grid don't move probabily meccanica problem or not Voltage on Motor or Motor broken

Log inor Register

to rate and post a comment

(148)
(19)

Tibor Ivanszky

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:12

Did you try another bucky? Most probably the solenoid which moves the grid is faulty. Contact me at tibor@ivanszky.com

Log inor Register

to rate and post a comment

(12)

Jerry Smith

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:12

E04 GEN None Arm fatal failure Generator detects
error in dialog with
arm.
Switch OFF Gantry, then switch ON.
If the problem continues, note ALL
messages and contact GEMS Field
Service.
E71 GAN Bucky Fail Bucky Failure Bucky cannot be
moved under normal
conditions.
1. Remove Bucky.
2. Reinstall Bucky.
3. Make a Rad Prep.
4. If failure continues, note ALL
messages, and contact GEMS Field
Service.
Exposures can be made without the
Bucky

Log inor Register

to rate and post a comment

(214)
(30)

CRAIG WAMBURA

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:12

I believe you have a grid movement issue.
118/012 is a limit of travel not found error caused by:
- There is abnormal friction. If the effort to produce movement is too high, the gain is artificially raised. An overcurrent detection is also possible.
- Too high gain and/or wrong offset
- defective photo-interrupter (F1) on 800-PL10
- Motor cannot move (open voice coil wires, defective electronic circuitry)
- Motor only moves in one direction - defective electronic circuitry (800-PL1)
- Optical end-of-travel cannot be reached before mechanical stop: 800-PL1 not centered.

Use Mammocom software to troubleshoot the grid and calibrate...send me results via whatsapp - +256782702243

Log inor Register

to rate and post a comment

(214)
(30)

CRAIG WAMBURA

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:12

027/002 means the grid start signal was not received or grid fault detected.

Log inor Register

to rate and post a comment

(327)
(64)
(2)

juan pablo amado

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:12

118/012 Grid movement: limit of travel not found

027/002 Grid sync: grid start signal not received or grid fault detected

Log inor Register

to rate and post a comment

(50)
(14)

tim cowell

re: GE DMR Plus Error E04 and E71 Assistance

April 16, 2019 10:13

027/002 Grid start signal not detected or grid fault detected. 118/012 Grid movement - limit of travel not found. You can check for grid motor voltage at 800 PL10 corner XJ2. you can check Hall Effect, on grid, approximately 8 ohms. Hall sensor is +4V(+/- 50mV)without any magnetic field measured on TP1/TP2(800-PL10)

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students.

 

advertisement

 

advertisement

 

advertisement

 

advertisement
对这个站点的通入和用途是受期限和条件我们支配 法律公告 & 保密性通知
物产和业主对 DOTmed.com,公司 Copyright ©2001-2019 DOTmed.com, Inc.
版权所有