Korvin Kel

CR 85-X

February 13, 2018 08:08

Hello.
Can someone help me with an explanation of the error SERVICE 24724? Tnx.

Log inor Register

to rate and post a comment

(130)
(9)

Ronald Swapp

re: CR 85-X

February 13, 2018 04:40

Error is:
E-IOC_ERACTL_OPENRELAIS
Reason: IOBUS control module: Erasure control board safety test:
Safety Relais does not open again.
Meaning: Safety relais on Erasure control board for interrupt power supply
to erasure lamps does not open.
Hardware Problem.
Cure: Replace Erasure control Board.

Send me an e-mail and I'll send you the service manual for the 85 and IMOS Show Error Viewer.
rswapp@comcast.net
Authorized AGFA Field Service Tech.

1 person liked this
Log inor Register

to rate and post a comment

(130)
(9)

Ronald Swapp

re: CR 85-X

February 13, 2018 04:40

My guess is you lost the Erasure Control Board, GS102 number: CM+9515564204.
Or you may have a problem with the Current Sense Board, GS104. part number: CM+9515564401.
Go for the GS102 First, I believe there is a micro relay that has a contact that is welded closed from over current.