Hi guys, i have a problem with Sterilclave 18B/24 - autoclave manufactured by COMINOX Italy.
The cycle start normally and work good till 25-30min, after that it show an error "hermetic alarm"
and the clave not continue working, just suspend here.
Can anyone help to resolve this problem. I can upload pictures or videos.
You should be having a 'INSUFFICIENT STEAM ALARM' indicating that the steam is leaking
But the chamber seal is hermetic meaning its airtight .... so you should be having an issue with your chamber seal.
Check your door gaskets...
clean internal disk on the door....remove any product residue....
Check the tilt of the sterilizer...(review installation manual)
If alarm still on....replace seal..
Contact for further instructions: craigwambura@outlook.com +256782702243 (WhatsApp)
Adepto Medical is an infusion pump rental, sales, and service company. Our focus is to deliver to our customers the best product and customer service support we can offer. Call (913) 261-9933
(1)
Borislav Ivanov
Sterilizer Sterilclave 18B/24
April 12, 2019 09:34
Hi guys, i have a problem with Sterilclave 18B/24 - autoclave manufactured by COMINOX Italy.
The cycle start normally and work good till 25-30min, after that it show an error "hermetic alarm"
and the clave not continue working, just suspend here.
Can anyone help to resolve this problem. I can upload pictures or videos.
to rate and post a comment
(47)
CRAIG WAMBURA
re: Sterilizer Sterilclave 18B/24
April 14, 2019 04:07
You should be having a 'INSUFFICIENT STEAM ALARM' indicating that the steam is leaking
But the chamber seal is hermetic meaning its airtight .... so you should be having an issue with your chamber seal.
Check your door gaskets...
clean internal disk on the door....remove any product residue....
Check the tilt of the sterilizer...(review installation manual)
If alarm still on....replace seal..
Contact for further instructions: craigwambura@outlook.com +256782702243 (WhatsApp)
to rate and post a comment
(6)
C. Michel Bessette
re: Sterilizer Sterilclave 18B/24
April 14, 2019 04:08
The most common fault with this style of autoclave is the door gasket not sealing.
C. Michel Bessette, RRT (Canada #72-659)
Servicios Tecnicos CMB7
16 Crescent
Kanata, Ontario. K2M 1L1
serviciostecnicoscmb7@gmail.com
Mobile: 0016138185321
to rate and post a comment
(1)
AMADI
re: Sterilizer Sterilclave 18B/24
April 14, 2019 04:08
first confirm that the thermostatic valve is working well
to rate and post a comment
(1)
AMADI
re: Sterilizer Sterilclave 18B/24
April 14, 2019 04:08
check the thermostatic valve it seems to be loosing pressure
to rate and post a comment
(1)
Borislav Ivanov
re: re: Sterilizer Sterilclave 18B/24
May 08, 2019 10:22