siemens arcadis tv-iris failure

(3)
(1)

garanich oleg

siemens arcadis tv-iris failure

July 20, 2021 10:50

Hello
siemens arcadis gives tv-iris failure ASP_E_7408 or
7409
sometimes after reboot everything is fine
Please tell me what to watch

Log inor Register

to rate and post a comment

(102)
(16)
(17)

Doscar Ernesto Perez Pereyra

re: siemens arcadis tv-iris failure

July 21, 2021 09:30

You need to enter to Error log for watch these error but normally fixed with the camera replace

Log inor Register

to rate and post a comment

(221)
(72)
(5)

Marko Shevchuk

re: siemens arcadis tv-iris failure

July 21, 2021 09:31

Hello!
7408
Explanation: a) TV-Iris failure.
TV-Iris: Unexpected movement
The TV-Iris movement was not activated, but the system detected a movement.
Action: a1) Check TV-Iris parameters.
a1) Check actual value potentiometer.
a2) Check wires.
a3) Check TV-Iris (I.I. optical system).
a4) The Main System D1 might be defect.

7409
Explanation: a) TV-Iris failure.
TV-Iris: Wrong position
If the failure 7400 to 7408 were detected,
this failure is displayed after each radiation release.
User warning:
The user must delete this failure
before radiation can be released again.
Action: a1) Check for the event log id with the source ASP_MSP 74*