Medilab Global provides high-quality pre-owned imaging equipment, parts, and service to the healthcare industry worldwide. Our number 1 position is to provide excellent customer service.. 305-234-0084 | Email: info@medilabglobal.com | Hablamos Español
Mohammed Abdullah
Planmeca proline PM2002
March 13, 2019 09:53
Hello All,
i have this machine gives error 61 in the stand by mode after power on,
if any one can help me with manuals and how can i fix this error,
thanks
with best regards
to rate and post a comment
(1)
Abd Ullah
re: Planmeca proline PM2002
March 14, 2019 04:26
Dear mohamed
This error code for processor communication
I can help you
abdlaamin0@gmail.com
to rate and post a comment
(3)
Matt Stiern
re: Planmeca proline PM2002
March 14, 2019 04:43
ERROR61:Generator Processor does not respond:
+33V below 24v
serial communication down
no power to generator processor bd
+5v shorted
Faulty SW Prom
Board defective
check cables and voltages disconnect motors and sensors
to rate and post a comment