Mark Lustig

Kodak PoC 140

August 28, 2013 03:03

Hello,

Kodak PoC 140.

After replacing rollers (front and back) the CR intermittently produces a blank film. The x-ray is fine. Our mR meter showed normal mR output during all our test exposures. The image begins to scan and you can tell it is "gray and grainy" then the processing takes over and blackens the film to a Q-35. Just like it does if it is a non-exposed plate. Again the is intermittent and does it about 1 out 5 exposures but sometimes it might not do it for 10 or 12 exposures. We replaced the nylon flex cable from the USB to the PMT. Didnt seem to make a difference.
Any thoughts out there?
Thanks,
Mark

Log inor Register

to rate and post a comment

(62)
(16)
(1)

JOSE HERRERA

re: Kodak PoC 140

August 28, 2013 05:31

Hello, please tell me which version of software has the POC, my email is jose.ig.herrera @ gmail.com. I think you have a software problem ..

Greetings.

José Herrera.

Log inor Register

to rate and post a comment

(5)

Duz

re: Kodak PoC 140

February 03, 2020 09:48

Hi Mark, Please how did you solve the problem? When cassette is exposed on my Kodak poc 140, no image is seen following scanning. What shows is only ghost/noise or a dark area at most. Please what can I do?