GE Revolution ACT 16 slices problem

October 22, 2019 10:41

Hi to all mates, i am new to this forum.
I am facing problem with Ge revolution ACT 16 slice CT, when daily warm-up is launched its aborted by CT, and the following errors displayed in errors journal.
Sat oct 19 16:48:32 2019
Host: tgp proc: tgp_subsys Error: 260100032
File: scTrigSeqActions. line: 1986
Sat Oct 19 16:48:32 2019
Host: Table Gantry Processor Ermes # : 260100032
Exception Class : Abort Severity: Primary
File : scTrigSeqActions.c Line# : 1986
Function : XRAY GENERATOR : Final Exposure Command
Scan Type : Axial Scan : 50272/1/2
Timeout on Exposure Command reset to X-ray off!
HW Error : Possible hardware problem.
Timeout is : 300 mS.
(KV: 80 mA: 20 GenOpt: 0x8 Freq: 984 Ax: 1000 Dur: 992).

Any help or support will be appreciated
Thanks in advance.



Log inor Register

to rate and post a comment

(49)
(13)

Brad

re: GE Revolution ACT 16 slices problem

October 23, 2019 10:25

Timeout on exposure command is the system has not been given command to expose. Possible issue with GUI or staff not hitting scan button in time and scan aborting?