Checkup failure Siemens Somatom Scope 2017

Tags: CT, SIEMENS

(1)
(1)

Marius Stoica

Checkup failure Siemens Somatom Scope 2017

November 16, 2021 01:56

Hello
I just encountered a problem when performing checkup procedure to a Somatom Scope Power from 2017. During the procedure is entering in a loop (repeating last scan until fatal error ). Any idea? Thank you in advance!

1 person liked this
Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

re: Checkup failure Siemens Somatom Scope 2017

November 17, 2021 10:11

Please check the event log. You can get details of the error. If it is aborting scan in warm up cycle then it could be either x-ray tube or the x-ray generator issues. Is aborting in filament adaptation cycle or defective channel correction or in the air calibration cycle?

Log inor Register

to rate and post a comment

(331)
(67)

Gabriel Cool

re: Checkup failure Siemens Somatom Scope 2017

November 22, 2021 11:26

When you have got this error? When system start filament adaptation or deffective channels? If first - then problem can be with tube. If second- check detector or irs

Log inor Register

to rate and post a comment