AGFA DX-G fails to power up

Tags: Radiology - R/F

(97)
(14)

Donald Bradbury

AGFA DX-G fails to power up

January 26, 2015 11:08

I have an AGFA DX-G CR that will fail to power up at turn on. No error messages. After several attempts system will finally boot up and operate fine. Any thoughts?

Log inor Register

to rate and post a comment

Lenny Place

re: AGFA DX-G fails to power up

January 27, 2015 10:13

verify your power supplies if they are good and within tolerance then you most likely will need to replace the drive and reload software (you could clone the drive which would prevent you from having to configure it)

Log inor Register

to rate and post a comment

(24)
(7)

Vo Quoc

re: AGFA DX-G fails to power up

January 27, 2015 10:32

HI Donald, the main PowerSupply will be die soon
you need to backup the main power supply for this CR
Here is the part code CM+9517098051 POWER SUPPLY (1 PIECE) (B616, B617)
Good Luck to you
Bao

Log inor Register

to rate and post a comment

(49)
(13)

Brad

re: AGFA DX-G fails to power up

January 27, 2015 11:06

Replace the both of the power supplies on the back.

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: re: AGFA DX-G fails to power up

February 06, 2015 04:16

What are the indicators other than failure to turn on? Will the output of either supply be low or non-existant? The manual shows the supplies to be 28 volt but no plus/minus perameters.

Log inor Register

to rate and post a comment